Dear Kamal,
the analog data outputs on 5890 II are located on the right top side of the GC (under a removable cover). There are usually three 12 pin rectrangle connectors marked Signal 1, Signal 2 and Remote. When specified in order, a connecting cable to Clarity A/D board equiped with those connectors is supplied. It can also be ordered separately. I am sending you a wiring diagram for this cable (in separate E-mail, this conference does not allow for enclosures), so you can prepare it or locate the corresponding wires on your integrator cable.
Please note, digital data acquisition from 5890 II is also available with the optional GC control module for Clarity, in case the GC is equiped with the RS232 communication port.
Best regards
Ivan Vins
DataApex
the analog data outputs on 5890 II are located on the right top side of the GC (under a removable cover). There are usually three 12 pin rectrangle connectors marked Signal 1, Signal 2 and Remote. When specified in order, a connecting cable to Clarity A/D board equiped with those connectors is supplied. It can also be ordered separately. I am sending you a wiring diagram for this cable (in separate E-mail, this conference does not allow for enclosures), so you can prepare it or locate the corresponding wires on your integrator cable.
Please note, digital data acquisition from 5890 II is also available with the optional GC control module for Clarity, in case the GC is equiped with the RS232 communication port.
Best regards
Ivan Vins
DataApex
Ivan Vinš
-
- Posts: 1
- Joined: Tue May 20, 2014 11:45 am
Re: HP 5890
Estimate: Ten years after this discussion, I encounter the same problem. I want to buy and use the Software clarity but they do not need to do so. I have a HP5890 series II without Rs232 via remote connected to HP3395 integrator , which is connected to the PC RS232 . I would not use the integrator and directly use the Clarity software. Could you please help me. Thank you so much
Martin
- Daniel Mentlik
- DataApex Support
- Posts: 354
- Joined: Fri Mar 27, 2009 3:15 pm
Re: HP 5890
Dear Martin,
thank you for your interest.
Generally you have two options:
a) get Clarity (or Clarity Lite) with Colibrick A/D Converter; this option will skip the integrator and get the data from the GC to the PC. You will still be forced to fill in the GC methods in the GC itself, but other than that you will be able to evaluate the data and synchronize the starts from the Clarity environment. For this solution you need:
For price estimate please contact Clarity distributor in your country.
Best Regards,
Daniel
thank you for your interest.
Generally you have two options:
a) get Clarity (or Clarity Lite) with Colibrick A/D Converter; this option will skip the integrator and get the data from the GC to the PC. You will still be forced to fill in the GC methods in the GC itself, but other than that you will be able to evaluate the data and synchronize the starts from the Clarity environment. For this solution you need:
- Clarity or Clarity Lite
- Colibrick A/D Converter (please specify how many detector signals you are using)
- Clarity
- GC Control Module
For price estimate please contact Clarity distributor in your country.
Best Regards,
Daniel
Daniel Mentlík
DataApex