Hi
The HPLC system is an agilent 1100 and the clarity version is 2.8.1.584
I setup my sequence and wnat to run it, but then he is waiting for the external start and doesn't start at all. when i remove the external start, clarity says there have to be an external start.
how shall i proceed?
Cheers
waiting for external start
-
- DataApex Support
- Posts: 10
- Joined: Fri Sep 04, 2009 4:10 pm
Re: waiting for external start
Hi,
to find out the reason we need to check the diagnostic files from your station, namely:
c:\clarity\clarity.cfg
c:\clarity\clarity.dsk
c:\clarity\systeminfo.txt
c:\clarity\log\ *.audit for the days the problem appeared
please send them to support@dataapex.com.
As a first guess I can recommend you to check the state of "Enable to start devices from Clarity" checkbox in Agilent 1100 Setup dialog, for proper function with autosampler it should be unchecked.
Jiri Kuntos
to find out the reason we need to check the diagnostic files from your station, namely:
c:\clarity\clarity.cfg
c:\clarity\clarity.dsk
c:\clarity\systeminfo.txt
c:\clarity\log\ *.audit for the days the problem appeared
please send them to support@dataapex.com.
As a first guess I can recommend you to check the state of "Enable to start devices from Clarity" checkbox in Agilent 1100 Setup dialog, for proper function with autosampler it should be unchecked.
Jiri Kuntos
Jiri Kuntos
-
- Posts: 2
- Joined: Wed Dec 05, 2012 12:27 pm
Re: waiting for external start
Hi
Thanks.
I send the files to the adress that you gave me. I also checked the box and it is unchecked. Is it right that this box is only in the VWD detector?
Cheers
Robert
Thanks.
I send the files to the adress that you gave me. I also checked the box and it is unchecked. Is it right that this box is only in the VWD detector?
Cheers
Robert
Robert Roos
Re: waiting for external start
Dear Robert,
yes, the Enable start from Clarity checkbox is present in the Agilent 1100 setup only for detectors (VWD in your case).
Note the Waiting for External Start probably indicates you are trying to run from the Single Analysis window - In Clarity for controlled autosamplers (like is your case) the Sequence window must be used (there is no field for the Vial number in the Single analysis window).
There is lot of checks performed by the sequence (valid vial position, injection volume etc.) which are dependent on the actually used method and which will be quite complicated to perform also in the Single analysis window.
yes, the Enable start from Clarity checkbox is present in the Agilent 1100 setup only for detectors (VWD in your case).
Note the Waiting for External Start probably indicates you are trying to run from the Single Analysis window - In Clarity for controlled autosamplers (like is your case) the Sequence window must be used (there is no field for the Vial number in the Single analysis window).
There is lot of checks performed by the sequence (valid vial position, injection volume etc.) which are dependent on the actually used method and which will be quite complicated to perform also in the Single analysis window.
Ivan Vinš