universal calibration

You can discuss anything concerning Clarity in this section.
Post Reply
leih01
Posts: 2
Joined: Wed Jan 30, 2013 9:09 am

universal calibration

Post by leih01 »

I want to use universal calibration. I try to set up Mark-Houwink constants for calibration points. However, even I input the constants in Calibration Options, the constants showing in the calibration table are always the default ones. How to set up the Mark-Houwink constants for the standards?
Hua Lei

User avatar
Ivan Vins
DataApex Support
DataApex Support
Posts: 183
Joined: Fri Mar 27, 2009 3:16 pm

Re: universal calibration

Post by Ivan Vins »

The Mark Houwink constants in the Calibration Options affect only the display of calibration curve in Calibration window, when no chromatogram is opened. For standards and unknowns, the constants need to be entered in the Sample Header (either before acquisition or on the calculations section in Result tab of the Chromatogram window. For imported chromatograms, the values entered in the Single Analysis window will be applied. The constants from chrommatogram will be applied to callibration when the respective level is calibrated.