Normalization method

You can discuss anything concerning Clarity in this section.
Post Reply
adedf
Posts: 7
Joined: Thu Jun 25, 2009 5:56 pm

Normalization method

Post by adedf »

Hi all,

Is it there in Clarity a provision to set up a normalization method using response factors ?

Best regards,

Ade
Ade

User avatar
Ivan Vins
DataApex Support
DataApex Support
Posts: 183
Joined: Fri Mar 27, 2009 3:16 pm

Re: Normalization method

Post by Ivan Vins »

The normalized% results will be displayed in the Amount% column of the Result table for chromatograms calculated by ESTD method, when the conditions for Norm% calculations are met. All peaks in the chromatogram must be integrated, identified and quantitated. (For peaks not calibrated, the response factor of a nearest homologue can be alternatively used).

To calculate results using response factors instead of calibration curve, enter the response factor values to the Resp. Factor column in the common part of the Calibration table and set the Curve fit type to Free.
Ivan Vinš

adedf
Posts: 7
Joined: Thu Jun 25, 2009 5:56 pm

Re: Normalization method

Post by adedf »

It works !!
Thank you
Ade