channel operation

You can discuss anything concerning Clarity in this section.
Post Reply
niranjan409
Posts: 1
Joined: Fri Mar 27, 2009 3:16 pm

channel operation

Post by niranjan409 »

i am using clarity 2.4.1.43 software which having facility of 2 channels 1 instrument.we have received 2 star/stop swiches. But my both the channels are getting operated by one of the switches only.Is it possible to operate 2 channels individually with two separate switches?
niranjan

User avatar
Petr Kohutek
DataApex Support
DataApex Support
Posts: 57
Joined: Fri Mar 27, 2009 3:15 pm
Location: Praha
Contact:

channel operation

Post by Petr Kohutek »

To operate the A/D channels independently you will have to purchase an add-on Instrument.

1 Instrument = 1 time base. This means that you are able to measure a chromatogram containing both signals, but you cannot use it for two independent chromatographs, where you need to measure independent analysis.

You can contact your dealer and order Add-on Instrument (o/n C55)

Petr Kohutek
DataApex

Petr Kohutek
DataApex