Implementation of Spark HPD selector valve as sampler

Posted: Tue Apr 09, 2019 4:58 pm
by Stellaschindler
I would like to make a software wish. I have a customer that has a Spark Holland ACE (dual clamp) and HPD (4 SSM). They use the HPD as a sampling pump and each SSM channel pumps 20 mL of sample onto the ACE cartridge. Right now, we have to write a Clarity method for each channel being used….which is 20 different methods.

Would DataApex be able to add a “Variable” solvent channel selection to HPD Actions Disp. Aspirate and Disp. Pump? There would also need to be a Setup Column in the Sequence so that the end user could designate the HPD channel to be used for each sample.

Internal Issue Tracking Number: 85437

Re: Implementation of Spark HPD selector valve as sampler

Posted: Fri May 10, 2019 11:36 am
by Ivan Vins
Dear Stella,

the suggestion is possible, but it will mean to write down a new control module to use the HPD SSM valves as an sampler, which will be rather laborious and costly. In case it is for one specific installation, I will suggest to consider using a VICI-Valco multiposition valve instead of the SSM valves, for it the control module is already available.

https://downloads.dataapex.com/document ... ampler.pdf