Control materials tracking in Clarity
About control materials (not calibrators), it could be useful to implement their tracking in the software (e.g. having the control materials being stated as Low Control and High Control in the sequence worklist and having their low/high range limits stored in the method, so that the end user can validate each analytical run just by verifying that the controls in the run do not excedeed those limits).
Gianluca Elli
- Daniel Mentlik
- DataApex Support
- Posts: 354
- Joined: Fri Mar 27, 2009 3:15 pm
Re: Control materials tracking in Clarity
Dear Gianluca,
can you please go into more details? I am quite not sure what exactly you would like to achieve, as Low Control and High Control does not ring a familiar note.
Thank you,
Daniel
can you please go into more details? I am quite not sure what exactly you would like to achieve, as Low Control and High Control does not ring a familiar note.
Thank you,
Daniel
Daniel Mentlík
DataApex
DataApex
Re: Control materials tracking in Clarity
In a clinical environment, after creating a calibration curve, users verify the calibration of the method by running (usually) a couple of control materials, a "low" one (LC, under the cutoff) and a high one (HC, over the cutoff).
These control materials are certified ones (lot nr. , expiry date...) with an average, desirable, value and a range of acceptability (eg. epinephrine high control: 100umol/L with an acceptability range from 92 to 108 umol/L). It would be nice to have at least the range of acceptability, for example as a column in the calibration table and having, in the sequence table, the possibility to set a sample as LC or HC. Should Clarity be also capable of checking the LC and HC result and warning the user when the result for any of them is out of the acceptability range, would be a strong plus.
These control materials are certified ones (lot nr. , expiry date...) with an average, desirable, value and a range of acceptability (eg. epinephrine high control: 100umol/L with an acceptability range from 92 to 108 umol/L). It would be nice to have at least the range of acceptability, for example as a column in the calibration table and having, in the sequence table, the possibility to set a sample as LC or HC. Should Clarity be also capable of checking the LC and HC result and warning the user when the result for any of them is out of the acceptability range, would be a strong plus.
Gianluca Elli
- Daniel Mentlik
- DataApex Support
- Posts: 354
- Joined: Fri Mar 27, 2009 3:15 pm
Re: Control materials tracking in Clarity
Dear Gianluca,
OK, now I understand. The desired behavior is in fact available in Clarity including the reaction on failed/passed test by using the SST Extension. How to set up the experiment:
On a chromatogram of such High Control standard, create a SST method in the Chromatogram window using the
To set up run to be checked by the SST method, set the sequence accordingly:
Would this procedure work for you?
OK, now I understand. The desired behavior is in fact available in Clarity including the reaction on failed/passed test by using the SST Extension. How to set up the experiment:
On a chromatogram of such High Control standard, create a SST method in the Chromatogram window using the
- SST - SST Result (to display the SST tab)
- SST - New (to create a new SST method)
- SST - Update from Calib (to load the list of peaks from a calibration file)
- Select the checkbox of the Tetrachlormethane in the table on the left
- Double-click the Amount Column in the table on the right to activate it
- Set the 1.1 value in the Lower Limit cell of the column, 1.3 into the Upper Limit cell.
- Right-click the table and select the SubParameters item. On it, switch to the Each Individual Value item and uncheck the %RSD Limit item (as you do not need it)
- You may hide the inactive rows and columns by using the SST - Show All Columns and SST - Show All Rows items in the menu
- Use the SST - Events command to open the dialog again.
- Set the tab as on next picture:
To set up run to be checked by the SST method, set the sequence accordingly:
- Right-click the sequence table and select the Setup Columns item
- Set the columns Open, Run Program, Program to Run, Parameters and Include in SST as visible (should be on the right side list)
- set the columns for the rows with High Control and Low Control samples as shown on the next picture:
Would this procedure work for you?
Daniel Mentlík
DataApex
DataApex
Re: Control materials tracking in Clarity
Hi Daniel, thank you.
Yes, the SST could be a workaround to have a quality control check on runs. We must think about including it in our offers to the customers, though, since it will be an additional cost....
Yes, the SST could be a workaround to have a quality control check on runs. We must think about including it in our offers to the customers, though, since it will be an additional cost....