Method should contain section specifying the Clarity layout

Posted: Thu Apr 08, 2010 11:11 am
by tony
Hi, Next update of clarity can you please add another page in the method chromatography setup, or add a new feature in the advanced tab that lets you put default parameters into the method for when the run has finished. ie settings for voltage, run time that are only applicable to that method. It is very frustrating when you are running numerous methods on 1 gc and have to alter these on the chromatograph window after the run has finished prior to printing to get a scaleable display. The current setup is only ideal if you are running 1 method!

Re: Method should contain section specifying the Clarity layout

Posted: Thu Apr 08, 2010 3:05 pm
by Ivan Vins
The graph and table preferences used for printing are stored in the desktop *.dsk file and could not be changed from a method.

In case the default settings for axes in Chromatogram window, Properties... (the Fixed axes not checked), which will autoscale the chromatogram to fit the window automatically, are not satisfactory, possible way will be to use the Launch Manager utility (introduced in Clarity version 2.8), which allows you to specify profiles to use specific configuration, project, desktop and method easily. You could then design numerous desktop files defining not only the graph axes ranges, but also the Result and Summary tables layout or method specific user column calculations.

By selecting the method profile from the Launch manager you will then open the corresponding project, method and desktop file by a single click.