Re-assigning internal standard peak

Posted: Mon Nov 27, 2017 5:20 pm
by TomaszK
Hi.

Is it possible to re-assign the internal standard peak using Clarity (v. 3.04.444)? My peaks shifted to the left over time and the instrument is now incorrectly assigning one of my analytes of interest as the internal standard.

Thanks in advance for any help.

Regards
Tomasz Kubczyk

Re: Re-assigning internal standard peak

Posted: Tue Nov 28, 2017 10:23 am
by Daniel Mentlik
Dear Tomasz,
assigning of any peak identifications is solely performed based on retention times as inserted in the calibration. To modify the identification, you need to change the retention time(s) and/or identification windows of compounds in the Calibration file you are using.

It may also be beneficial for you to use the Reference Peak functionality - for more details please see the documentation for it: Using a reference peak to improve compound identification

Version you are using is quite old though, we would recommend updating to a new version which is free of charge. Please just check the compatibility of the OS and hardware prior performing the update.

Re: Re-assigning internal standard peak

Posted: Wed Nov 29, 2017 11:46 am
by TomaszK
Hi Daniel

Thank you very much for your answer. It has been very useful.

Regards
Tom