Input over range

You can discuss anything concerning Clarity in this section.
Post Reply
adedf
Posts: 7
Joined: Thu Jun 25, 2009 5:56 pm

Input over range

Post by adedf »

Dear All,

Is it possible to report a warning if , during an analysis run, the input voltage run over the maximum (i.e. 1250 mv)?
Best regards,
Ade

User avatar
Daniel Mentlik
DataApex Support
DataApex Support
Posts: 353
Joined: Fri Mar 27, 2009 3:15 pm

Re: Input over range

Post by Daniel Mentlik »

Dear Ademaro,
first of all let me apologize for the delay - we have moved our server to another hosting and your topic came in the middle of the operation, thus took some time to recover from backups and synchronize it again.

As for the question, yes, such operation is possible using the Event table in Clarity. Set up the row of the table as shown on the screenshot:
forum-input-over.png
In this case, Clarity would cause a beep.exe file (available in Utils directory) to be run when signal level will rise over 1249 mV, giving a sound warning. Of course you can set more types of the output action, up to the stopping acquisition (Output Type Command, then select from available options in Output column).

Best Regards,
Daniel
Daniel Mentlí­k
DataApex

Post Reply