Unidentified peaks response factor
Posted: Wed May 09, 2012 3:30 pm
by adedf
Dear all,
Is it possible to enter, instead of the value, a pointer to a known component in the calibration table?
This will allow the automatic recalibration for the unknown peaks as well.
Best regards
Re: Unidentified peaks response factor
Posted: Thu May 10, 2012 11:23 am
by Daniel Mentlik
Dear Ademaro,
please could you describe in more detail how exactly do you want to use the calculations of unidentified peaks? I am sure we can come up with a solution.
Aside from that, response factor for non-identified peaks can be set on two places in Clarity. The first one is in the section
Undefined Peaks on the
Method Setup - Calculation tab, and the value entered here will be used as a response for any peak not present in the calibration file resulting from the given method. The value can be later changed for particular chromatogram from the
Chromatogram window,
Results tab, right section.
The other way how to set custom responses for unidentified peaks, this time individual to particular peaks, is by using the
Free calibration option. Please check the chapter 3.6.2 in the Clarity Reference Guide (for download on our web pages,
here) for detailed description of the function and the types of calculations in Clarity.
Best Regards,
Daniel
Re: Unidentified peaks response factor
Posted: Fri May 11, 2012 5:21 pm
by adedf
Dear Daniel,
The scope is quantify the peaks not in the calibration table by assigning the response factor of a peak whose RF is in the calibration table.
Example: In a chromatogram you may have many peaks, not listed in the calibration table, and you want these peaks reported as they were n-hexane ( that is one of the calibrated peaks).
I understand that Clarity let me assign a RF, but not the RF of a peak in the calibration table.
Is that correct ? Is there any workaround?
best regards
Re: Unidentified peaks response factor
Posted: Fri May 11, 2012 5:28 pm
by Ivan Vins
Dear Ademaro,
there are two ways to solve this:
1) Manually copy the hexane response factor from Calibration to the Calculation tab, unidentified peak response.
2) Add the unknown peaks to the calibration table and assign them the hexane response factor using the Calculate by column (hidden by default, use Setup columns in the Calibration table).
3) In case there is variable number and positions of such peaks, you could assign them as a group and calibrate it in the above suggested way.
Best regard
Ivan
Re: Unidentified peaks response factor
Posted: Mon May 14, 2012 5:27 pm
by adedf
Dear Ivan,
The solution 3) is the viable one.
Unfortunately I'm not able to implement it.
I can implement solution 2)as described, but applying it to a peak group shows an error on the "peak type" column.
I'll send you the screenshot by mail.
Can you help?
Best regards
Re: Unidentified peaks response factor
Posted: Thu May 17, 2012 8:08 am
by Ivan Vins
Dear Ademaro,
sorry for the wrong info, the Calculate By option is not implemented for Groups. Will be fixed in some future version.
Ivan Vins