I would like to ask is it possible to use Amount parameter in SST modul if I use calibration file that defines group of peaks.
For my customer it would be helpful to determine by SST test if the Amount results is whithin of their specification sepcification. They define group of peaks and calibrate it by ESTD.
Best regards,
JCzajka
Amount parameter in SST modul using group calibration file
-
- Posts: 16
- Joined: Fri Jun 12, 2009 9:49 am
Amount parameter in SST modul using group calibration file
Jarosław Czajka
Re: Amount parameter in SST modul using group calibration file
The optional SST module allows you to calculate average and RSD values and also check, if they are within preset limits. The parameters include area, height, amount and their percentages as well as performance values like efficiency, symmetry and resolution) . The calculations are made for compounds identified by the calibration file, there is no difference if the compound is ordinary peak or peak group.