Posted: Thu Sep 21, 2006 5:55 pm
by techlab
Does anyone have experience with the universal pump control module to
build a serial controller for gradient pumps ? If yes, which pumps
can be controlled?
Mit freundlichen Grüßen,
Werner Röpke
************************************
Techlab GmbH
Werner Röpke
Asseblick 4
D-38173 Erkerode
Tel: 05305/930203
Fax: 05305/930208
roepke@techlab.de
www.techlab.de
Amtsgericht Braunschweig HRB 7566
************************************
Pump control via RS232
Posted: Thu Sep 21, 2006 5:55 pm
by Ivan Vins
Dear Werner,
To use a UNI LC Pump control module, the pump should have a RS232 serial line communication with a simple ASCII command communication protocol available. The pump should have a command for setting the actual flow and composition, the gradient is formed in Clarity and sent as immediate commands to the pump. Either high pressure gradients or low pressure gradients can be controlled. With high pressure gradients it may be necessary to use independent drivers (separate com ports) for each pump (LabAlliance), however some pumps allow for addressing several of them on the same serial line (Flom, Kontron with multiport interface). More details you can find in the LC UNI pump control module manual available for download from our web pages.
Profiles for LabAlliance Series II, Series III, Prep 100 and 1500 pumps, Knauer Smartline 1000, Ecom Alpha and Beta are available, others are available for testing.
Best regards
Ivan Vins
DataApex, Ltd
Podohradska 1
15500 Praha 5
Czech Republic
Phone +420 251612067
Fax +420 251613507
www.dataapex.com
Pump control via RS232
Posted: Thu Sep 21, 2006 5:55 pm
by techlab
Dear Ivan,
does anyone use older Knauer K-1000 series?
Is the Merck-Merck-Hitachi Gradient pump controllable via RS232?
Werner
Mit freundlichen Grüßen,
Werner Röpke
************************************
Techlab GmbH
Werner Röpke
Asseblick 4
D-38173 Erkerode
Tel: 05305/930203
Fax: 05305/930208
roepke@techlab.de
www.techlab.de
Amtsgericht Braunschweig HRB 7566
************************************
Pump control via RS232
Posted: Thu Sep 21, 2006 5:55 pm
by Ivan Vins
Dear Werner
Knauer K-1000 does not support the simple ASCII protocol, it could not be controlled by Clarity UNI LC pump control module. For K-500 the control should be available. However Knauer is planning to develop drivers for his older pumps in the future too.
I am not familiar enough with Merck-Hitachi pumps, in case they support some ASCII communication, we do not have the protocols.
Best regards
Ivan Vins