Dear Sir
We have been working with a customer for an order for Clarity.
He wants us to demonstrate Resolution in the peak reports. Unfortunately we
are not able to find the feature in the demo software.
Please guide us on how should we show him that our software gives all the
features.
Also give us a demo CD for SST also, as the same customer is planning to buy
the SST also after we show him the demo.
Thanks\
RadhaKrishnan
Sales Engineer
www.ais-india.com
Internal Virus Database is out-of-date.
Checked by AVG Anti-Virus.
Version: 7.0.289 / Virus Database: 265.4.6 - Release Date: 12/5/2004
Resolution in the peak reports- Clarity- Order against Demo
-
- Posts: 1
- Joined: Fri Mar 27, 2009 3:16 pm
Resolution in the peak reports- Clarity- Order against Demo
Radha Krishnan
Resolution in the peak reports- Clarity- Order against Demo
Dear RadhaKrishnan,
the Resolution (as other special parameters) you can find at a special
tab "Performance" in Chromatogram window.
This tab is hidden by default. To display go to the menu Result -
Performance Table.
To display Resolution column choose "From width at 50%" in section
Column calculation in right part of Performance Tab.
The Resolution tab is the last one in displayed table.
By "Drag and Drop" you can easily move this column at the beginning of
the table.
By command Setup Columns from local menu of this table you can rearrange
the table and hide unused columns.
The Resolution parameter is a part of SST module as well.
The SST (System Suitability Test) module is an integral part of each
Clarity demo version.
Click on the SST tab in Chromatogram window.
By command Open... from local menu display a sample of SST parameters.
For detailed information please use Clarity on-line help or download
comprehensive SST manual from our web site.
Jan Hruby
the Resolution (as other special parameters) you can find at a special
tab "Performance" in Chromatogram window.
This tab is hidden by default. To display go to the menu Result -
Performance Table.
To display Resolution column choose "From width at 50%" in section
Column calculation in right part of Performance Tab.
The Resolution tab is the last one in displayed table.
By "Drag and Drop" you can easily move this column at the beginning of
the table.
By command Setup Columns from local menu of this table you can rearrange
the table and hide unused columns.
The Resolution parameter is a part of SST module as well.
The SST (System Suitability Test) module is an integral part of each
Clarity demo version.
Click on the SST tab in Chromatogram window.
By command Open... from local menu display a sample of SST parameters.
For detailed information please use Clarity on-line help or download
comprehensive SST manual from our web site.
Jan Hruby
Jan Hrubý