I'm fairly new to Clarity VA as well as this forum.
I have following error atm:
'Chromaster CM5160 LC 1 (SN...): SetupTimeProgram: Parameter Error (CM 51x0 Pump)'
I encountered this error, multiple times already, when trying to 'Purge...' the tubings of our HPLC system with solvents. They are regular solvents we use.
It's probably something stupid, but it would be nice if I can prevent this in the future
this is an error introduced in ClarityVA version 7.4.0.75, fixed in the current release 7.4.1.88.
Due to a bug in the control dll, the 5160 pump wrongly rejects maximum pressure set over 40 MPa.
Update is available directly from ClarityVA (when connected to internet) or through Hitachi approved ClarityVA distributors.
this is an error introduced in ClarityVA version 7.4.0.75, fixed in the current release 7.4.1.88.
Due to a bug in the control dll, the 5160 pump wrongly rejects maximum pressure set over 40 MPa.
Update is available directly from ClarityVA (when connected to internet) or through Hitachi approved ClarityVA distributors.
C:\ClarityVA\CFG\clarityVA.cfg
C:\ClarityVA\CFG\clarityVA.dsk
C:\ClarityVA\CFG\systeminfo.txt
C:\ClarityVA\CFG\others.ini
C:\ClarityVA\CFG\DebugLogs\badtrace.txt and all *.dmp files (may not be present, they are created during some problems only)
C:\ClarityVA\CFG\Audit_Trails\ *.audit for the days the problem appeared
files related to the problem located in the C:\ClarityVA\DataFiles
•method (*.met)
and communication record created after adding a section to
C:\ClarityVA\CFG\clarityVA.cfg
C:\ClarityVA\CFG\clarityVA.dsk
C:\ClarityVA\CFG\systeminfo.txt
C:\ClarityVA\CFG\others.ini
C:\ClarityVA\CFG\DebugLogs\badtrace.txt and all *.dmp files (may not be present, they are created during some problems only)
C:\ClarityVA\CFG\Audit_Trails\ *.audit for the days the problem appeared
files related to the problem located in the C:\ClarityVA\DataFiles
•method (*.met)
and communication record created after adding a section to
Note the error is rather generic when sending the pump time program fails for any reason.
Another possibility will be disabling the solvents in LC Gradient options - Clarity sends parameters only for enabled solvents, in case the pump stores non-zero values for the disabled ones, total may exceed 100%.