Baseline "OVER" during equilibrating

You can discuss anything concerning Clarity in this section.
Post Reply
ACSolutions
Posts: 11
Joined: Wed Nov 18, 2015 12:33 am

Baseline "OVER" during equilibrating

Post by ACSolutions »

Before analysis, when checking if the baseline is stable, we only see a flat line and the message "OVER" on the bottom for all 4 of the signals.

When running a sample, the baseline gives a normal image (fluctuations, noise, peaks) (image included). After analysis, the acquisition window again shows "OVER". AUTOZERO on the detector does not change.

Configuration: Chromaster with DAD
Attachments
image for the "OVER" msg in acquisition window during equilibration
image for the "OVER" msg in acquisition window during equilibration
kirsten de keyser

User avatar
Daniel Mentlik
DataApex Support
DataApex Support
Posts: 354
Joined: Fri Mar 27, 2009 3:15 pm

Re: Baseline "OVER" during equilibrating

Post by Daniel Mentlik »

Dear Kirsten,
to understand the reason of the problem, we need to know more details - the situation looks really weird.
The best way would be a TeamViewer session if the PC is connected to the Internet - please send us the email to support@dataapex.com and we will schedule a session. If the PC is not connected to the Internet, we will suggest further diagnostics over email.

Best Regards,
Daniel
Daniel Mentlí­k
DataApex

ACSolutions
Posts: 11
Joined: Wed Nov 18, 2015 12:33 am

Re: Baseline "OVER" during equilibrating

Post by ACSolutions »

Hi Daniel,

The PC is not connected to the internet. I will inform the VWR Technician, because she will have to solve this. I would have been of help if it was just a simple software setting, but if it is more complicated it's best VWR takes this in their own hands.
kirsten de keyser

User avatar
Ivan Vins
DataApex Support
DataApex Support
Posts: 183
Joined: Fri Mar 27, 2009 3:16 pm

Re: Baseline "OVER" during equilibrating

Post by Ivan Vins »

Dear Kirsten,

in fact you observe a normal behavior for the Chromaster DAD - note the detector is not giving any signal data outside run, the control module is just providing simulated out of range signal for the DataAcqusition window (needed to synchronize in case there will some other signals).

To evaluate the baseline before run you need to use the Check Signal button in Device Monitor - it will invoke a run in the detector, but not start run in Clarity. It need to be switched off before starting the run.
Ivan Vinš