Search found 166 matches

by Ivan Vins
Mon Aug 05, 2013 12:19 pm
Forum: Discussing Clarity
Topic: Automatic run?
Replies: 3
Views: 1856

Re: Automatic run?

Dear Domenico, note the command line parameter Delayed_exit, which will cause Clarity to exit after the period of time specified in seconds if it has no running acquisition. This will allow you to close Clarity after the sequence is finished. To start next sequence you will need to prepare a batch f...
by Ivan Vins
Mon Jul 15, 2013 11:39 am
Forum: Discussing Clarity
Topic: Automatic run?
Replies: 3
Views: 1856

Re: Automatic run?

Dear fenmick, I will suggest to use a simple batch file using command line parameters. For details see chapter 10.2 in Clarity Reference Guide http://downloads.dataapex.com/documentation/clarity/manuals/solutions/basic-clarity-reference-guide.pdf to open Clarity, send the respective method and perfo...
by Ivan Vins
Wed Jul 03, 2013 8:43 am
Forum: Discussing Clarity
Topic: saturated peak marks in result table
Replies: 6
Views: 3166

Re: saturated peak marks in result table

Dear Sunil Jadhav, for this purpose I will reccomend to use the optional SST extension http://downloads.dataapex.com/documentation/clarity/manuals/extensions/extensions-sst.pdf which allows you to compare some peak parameters (including Peak Height) with predefined limits and report the Pass/Fail re...
by Ivan Vins
Fri Jun 28, 2013 9:24 am
Forum: Discussing Clarity
Topic: Unexpected STOPPING state (YL9100 and Agilent 1200)
Replies: 1
Views: 1147

Re: Unexpected STOPPING state (YL9100 and Agilent 1200)

Dear lilylara, as the post is off topic for this thread, we will move it to separate thread. Stopping state occurs typically when data expected by Clarity from detector were not received. In your configuration it is most probably caused by the System, Configuration, Settings, where the selected opti...
by Ivan Vins
Mon Jun 24, 2013 6:02 pm
Forum: Discussing Clarity
Topic: Agilent 1100, uWPS (G1377A), CapPump(1376A) non-LAN control
Replies: 1
Views: 1156

Re: Agilent 1100, uWPS (G1377A), CapPump(1376A) non-LAN control

Dear Ben, the info on web is corrrect - there are two control modules for Agilent 1100/1200 systems available: older one (no further developed) Agilent 1100/1200 control, based on the communication protocols for individual modules as obtained from Agilent. It supports just a limited selection of mod...
by Ivan Vins
Wed Jun 05, 2013 11:21 am
Forum: Discussing Clarity
Topic: Sample Amount - Sequence Table
Replies: 1
Views: 1159

Re: Sample Amount - Sequence Table

Dear Marco, this is designed behaviour - when the Sample Amount is set to non zero value, it will replace the calculated total in Amount (unit) column to calculate the Amount(%) column. Refer to the HELP or manuals for the respective calculations used. You could however create an user column to perf...
by Ivan Vins
Thu Jan 31, 2013 3:12 pm
Forum: Discussing Clarity
Topic: flow marker
Replies: 1
Views: 1104

Re: flow marker

The collumn is hidden in the default settings, you need to reveal it by using the Setup columns from the Result table contect menu (right click in the table area)
by Ivan Vins
Thu Jan 31, 2013 3:10 pm
Forum: Discussing Clarity
Topic: universal calibration
Replies: 1
Views: 1084

Re: universal calibration

The Mark Houwink constants in the Calibration Options affect only the display of calibration curve in Calibration window, when no chromatogram is opened. For standards and unknowns, the constants need to be entered in the Sample Header (either before acquisition or on the calculations section in Res...
by Ivan Vins
Wed Jan 02, 2013 5:57 pm
Forum: Discussing Clarity
Topic: waiting for external start
Replies: 3
Views: 1887

Re: waiting for external start

Dear Robert, yes, the Enable start from Clarity checkbox is present in the Agilent 1100 setup only for detectors (VWD in your case). Note the Waiting for External Start probably indicates you are trying to run from the Single Analysis window - In Clarity for controlled autosamplers (like is your cas...
by Ivan Vins
Tue Nov 13, 2012 8:11 pm
Forum: Discussing Clarity
Topic: yl clarity sampling frequency setting of yl9120
Replies: 3
Views: 2183

Re: yl clarity sampling frequency setting of yl9120

Dear Laszlo,

please could you send us the acquired chromatogram to support@dataapex.com? I could not check the method file without corresponding configuration, it is however recorded in the chromatogram.