Search found 166 matches

by Ivan Vins
Wed Nov 19, 2014 7:35 am
Forum: Discussing Clarity
Topic: Long analysis with multiple methods
Replies: 3
Views: 2387

Re: Long analysis with multiple methods

The above application was described in some detail https://theanalyticalscientist.com/app-notes/0014/automated-sw-solution-for-preparative-hplc/ The methods could be run using the Sequence table. The sequence will perform the methods automatically in order as specified in the table, starting a next ...
by Ivan Vins
Wed Nov 05, 2014 6:20 pm
Forum: Discussing Clarity
Topic: pda functions
Replies: 3
Views: 2471

Re: pda functions

Dear Samira, please provide us some info: Your Clarity station serial number (in the main window Help menu, About box, S/N xxx-yyyyy) and version Manufacturer and model of your PDA detector Preferably please send the chromatogram as *.prm file as suggested in previous post, the above info is contain...
by Ivan Vins
Mon Aug 25, 2014 6:08 pm
Forum: Discussing Clarity
Topic: Cut start of a chromatogram
Replies: 3
Views: 2025

Re: Cut start of a chromatogram

Dear Carsten, 1) any reason to shift the retention times at all? As long as the time shift will be reproducible, it does not influence anything. 2) it is possible to perform some events before the run start, so it may be possible to synchronize start acquisition with turning of the valve - however i...
by Ivan Vins
Fri Aug 01, 2014 11:53 am
Forum: Discussing Clarity
Topic: Noise and Drift in nanoRIU units
Replies: 4
Views: 2253

Re: Noise and Drift in nanoRIU units

Dear Rick, the noise and drift are generally dispayed with units used for the respective signal. The solution of your problem will depend on the way the signal is acquired: Signals aquired from the detector analog output through the A/D board have default units mV, they could be amended in the Syste...
by Ivan Vins
Mon Jul 28, 2014 10:06 am
Forum: Discussing Clarity
Topic: How to determine the purity of a peak
Replies: 3
Views: 2111

Re: How to determine the purity of a peak

1) The peak purity profile for selected peak is displayed in the Peak Purity View pane of the PDA Chromatogram window. 2) Absorbance Threshold is used in the same way as Peak Purity Threshold to eliminate spectra with values below those limits (typically containing just noise) from the Peak Purity e...
by Ivan Vins
Fri Jul 18, 2014 4:35 pm
Forum: Discussing Clarity
Topic: How to determine the purity of a peak
Replies: 3
Views: 2111

Re: How to determine the purity of a peak

Peak Purity value is calculated as a match between the peak apex spectrum and all the spectra thorough the peak boundaries. Perfect match is represented by 1000. The values are plotted in the Peak profile window, the average is reported in the Result table. For pure peaks, the profile should be more...
by Ivan Vins
Wed Jun 25, 2014 6:31 pm
Forum: Discussing Clarity
Topic: Automatic shutdown on Clarity with Agilent HPLC
Replies: 1
Views: 1831

Re: Automatic shutdown on Clarity with Agilent HPLC

Recommended approach for system shutdown after end of sequence is to create a shutdown method (with lamp off and stopped pumps) to be run as a last one in the sequence. However, the Agilent ICF control does not allow to switch lamps off from method (this was possible in the older Agilent 1100/1200 c...
by Ivan Vins
Mon Jun 16, 2014 11:33 am
Forum: Discussing Clarity
Topic: saturated peak marks in result table
Replies: 6
Views: 3199

Re: saturated peak marks in result table

Dear Sunil,

for this purpose the NGA extension is recomended, it is designed to calculate the calorific values according to the commonly used norms and enables to combine signals from multiple detectors

http://www.dataapex.com/product.php?id= ... ns-nga.php

Best regards
by Ivan Vins
Mon Apr 28, 2014 12:48 pm
Forum: Discussing Clarity
Topic: pda functions
Replies: 3
Views: 2471

Re: pda functions

Dear Samira, my guess is the acquired chromatogram is not containing the spectral data for some reason (spectral detector not configured to instrument, spectral data acquisition not enabled in method, ...) Please copy the respective chromatogram as *.prm file to flash disc and send it from some comp...
by Ivan Vins
Fri Mar 21, 2014 10:30 am
Forum: Discussing Clarity
Topic: creating and matching library to unknown compouds
Replies: 7
Views: 3529

Re: creating and matching library to unknown compouds

Dear Samira,

in the PDA window set the Spectral View and Spectral Library View.

In Spectral Library view you will see the table with spectra in currently opened library (Use Library, Open), the spectra could be displayed in the Spectral view by the Show checkbox