Search found 166 matches

by Ivan Vins
Wed Oct 12, 2005 3:36 pm
Forum: Discussing Clarity
Topic: Data apex software for use with Agilent 1050 and 1100.
Replies: 1
Views: 2100

Data apex software for use with Agilent 1050 and 1100.

Dear John. the 1100 series LC can be controlled by Clarity (pumps, autosamplers, thermostats and detectors) with digital data acquisition, including the spectra from DAD and FLD detectors. HP-IB communication is used, the LAN communication is in development. We do not plan to add the control for 105...
by Ivan Vins
Fri Oct 07, 2005 4:18 pm
Forum: Discussing Clarity
Topic: Running a sequence without AS
Replies: 1
Views: 2181

Running a sequence without AS

Dear Arnau, there is not a problem to run a sequence in Clarity without an autosampler. Even passive sequence will send the method to controlled instruments when a new line is activated. With active sequence, the Ready out assigned digital output is activated (its state changed) - for not controlled...
by Ivan Vins
Mon Sep 26, 2005 4:02 pm
Forum: Discussing Clarity
Topic: Signal "ready" for AS
Replies: 1
Views: 2212

Signal "ready" for AS

Dear Sir, When using Active sequence in Clarity, the level of the Ready Out digital output signal is changed by the Sequence, when a next injection should be performed. This signal should be connected to the AS Start input, the appropriate Ready out wire is set in the System configuration window.. W...
by Ivan Vins
Sun Sep 04, 2005 5:49 pm
Forum: Discussing Clarity
Topic: Knauer detectors
Replies: 1
Views: 2030

Knauer detectors

Dear Rui, the digital data acquisition from detectors is our high priority. We are working on development of the necessary drivers, at present time the digital data acquisition is available for Agilent 5890/68xx GC and 1100 series LC detectors. We hope to have the possibility for digital data acquis...
by Ivan Vins
Mon Aug 08, 2005 9:58 pm
Forum: Discussing Clarity
Topic: UV vis spectrophotometer as and hPLC detector with Clarity.
Replies: 2
Views: 2460

UV vis spectrophotometer as and hPLC detector with Clarity.

Dear Carlos, we are considering the development of an universal detector driver (similar to the UNI pump control module), which will enable the digital data acquisition over RS232 serial line from devices with a simple ASCII protocols available. This could solve the described situation with your UV-...
by Ivan Vins
Sat May 21, 2005 5:02 pm
Forum: Discussing Clarity
Topic: Signal/Noise Calculation
Replies: 4
Views: 3575

Signal/Noise Calculation

Dear Mr. Sivamaran. this can be easily achieved by using the User columns calculations. 1) determine the peak width of peak of interest - the W05 value is reported in results table for UNCAL type of calculation. 2) set the interval for noise determination in the blank chromatogram - use the Integrat...
by Ivan Vins
Sat May 21, 2005 5:02 pm
Forum: Discussing Clarity
Topic: Signal/Noise Calculation
Replies: 4
Views: 3575

Signal/Noise Calculation

<t>Dear Mr. Sivamaran,<br/> <br/> Similar bug occured in certain Clarity 2.3 versions, the bipolar settings of the detector were errorneously identified as unipolar thus the baseline was shifted by a half of the used range and the signal height was reduced to half. Upgrade to current version availab...
by Ivan Vins
Thu May 12, 2005 10:16 am
Forum: Discussing Clarity
Topic: Measure and calculate breakthrough time
Replies: 1
Views: 1994

Measure and calculate breakthrough time

Dear Ingo, the calculation can be performed by using the user columns in result table, as there are variables in the sample header (Sample amount, Dilution, ISTD amount) which can be used for the calculations. The problem will be with the determination of the breakthrough time - the only suggestion ...
by Ivan Vins
Sat May 07, 2005 7:51 pm
Forum: Discussing Clarity
Topic: Validation on Clarity
Replies: 1
Views: 1971

Validation on Clarity

<t>Dear Sivamaran.<br/> <br/> the description of the data files as well as the limiting values for their evaluation are described in the Clarity Validation Kit manual. You can download it from our web pages, section downloads, manuals in a pdf format.<br/> <br/> Best regards<br/> <br/> Ivan Vins<br/...
by Ivan Vins
Fri Apr 08, 2005 4:03 pm
Forum: Discussing Clarity
Topic: omitting the A/D converter
Replies: 1
Views: 2062

omitting the A/D converter

Dear Jan, the immediately available solution is to use our Clarity Eval version for importing the data from your PC program as ASCII text files. To be able to acquire the data directly to Clarity a development of a driver will be necessary (we have digital data acquisition from Agilent 6890, 5890 GC...