- Fri Feb 09, 2018 1:27 pm
- Forum: Discussing Clarity
- Topic: error 2000 Setup time program
- Replies: 6
- Views: 7069
Re: error 2000 Setup time program
Dear Evelyn, in such case we will need the diagnostic files C:\ClarityVA\CFG\clarityVA.cfg C:\ClarityVA\CFG\clarityVA.dsk C:\ClarityVA\CFG\systeminfo.txt C:\ClarityVA\CFG\others.ini C:\ClarityVA\CFG\DebugLogs\badtrace.txt and all *.dmp files (may not be present, they are created during some problem...
- Fri Feb 09, 2018 1:03 pm
- Forum: Discussing Clarity
- Topic: error 2000 Setup time program
- Replies: 6
- Views: 7069
Re: error 2000 Setup time program
Hmm... I have done an upgrade of the software to version 7.4.1.95 (which is the most recent). But I still get the error...
- Fri Feb 09, 2018 12:50 pm
- Forum: Discussing Clarity
- Topic: error 2000 Setup time program
- Replies: 6
- Views: 7069
Re: error 2000 Setup time program
Dear Evelyne, this is an error introduced in ClarityVA version 7.4.0.75, fixed in the current release 7.4.1.88. Due to a bug in the control dll, the 5160 pump wrongly rejects maximum pressure set over 40 MPa. Update is available directly from ClarityVA (when connected to internet) or through Hitach...
- Fri Feb 09, 2018 12:07 pm
- Forum: Discussing Clarity
- Topic: error 2000 Setup time program
- Replies: 6
- Views: 7069
error 2000 Setup time program
Dear all, I'm fairly new to Clarity VA as well as this forum. I have following error atm: 'Chromaster CM5160 LC 1 (SN...): SetupTimeProgram: Parameter Error (CM 51x0 Pump)' I encountered this error, multiple times already, when trying to 'Purge...' the tubings of our HPLC system with solvents. They ...